Electronic ph meter
- gregorach
- Under the Table
- Posts: 1912
- Joined: Wed Jun 08, 2011 10:07 am
- Location: Edinburgh
- Contact:
Re: Electronic ph meter
Yeah, I've become extremely sceptical about the value of pH strips since buying a decent meter.
When calibrating, bear in mind that the buffers are surprisingly sensitive to temperature...
When calibrating, bear in mind that the buffers are surprisingly sensitive to temperature...
Cheers
Dunc
Dunc
-
- Even further under the Table
- Posts: 2514
- Joined: Sun Apr 01, 2012 5:38 pm
- Location: Wirral, Merseyside
Re: Electronic ph meter
Does anyone know where to get buffer solution at a reasonable price? Mine cost just short of a tenner for a 250 ml 7,4 pair and they don't tend to last very long.
Best wishes
Dave
Dave
Re: Electronic ph meter
My ph meter has thermometer also so I guessed it was factoring in the temp at all points...?gregorach wrote:Yeah, I've become extremely sceptical about the value of pH strips since buying a decent meter.
When calibrating, bear in mind that the buffers are surprisingly sensitive to temperature...
- gregorach
- Under the Table
- Posts: 1912
- Joined: Wed Jun 08, 2011 10:07 am
- Location: Edinburgh
- Contact:
Re: Electronic ph meter
Yeah, but the buffers themselves are only accurate at the correct temperature, and they're surprisingly sensitive to variations. Your typical 7.01 pH buffer is only actually 7.01 pH at exactly 25 degrees (or whatever it states on the bottle). The temperature correction for the meter is an entirely separate matter. Calibration must be performed at the correct temperature. Usually this requires placing the containers holding the buffers into a water bath - ideally thermostatically controlled, but you can just about get away with adjusting the temperature with hot or cold water until your buffers are at their correct temperatures. Only then can you proceed to calibrate the meter - otherwise, you're actually de-calibrating it.
Cheers
Dunc
Dunc
Re: Electronic ph meter
How inaccurate have you found the PH strips to be, Dunc?gregorach wrote:Yeah, I've become extremely sceptical about the value of pH strips since buying a decent meter.
Re: Electronic ph meter
Flipping heck. The manual didnt say that. The other thing was it said it recognises the 7.0ph solution on calibration. I then thought it would measure temp and calibrate it knowing the buffer temp versus accurate ph7.0. That is clearly possible....gregorach wrote:Yeah, but the buffers themselves are only accurate at the correct temperature, and they're surprisingly sensitive to variations. Your typical 7.01 pH buffer is only actually 7.01 pH at exactly 25 degrees (or whatever it states on the bottle). The temperature correction for the meter is an entirely separate matter. Calibration must be performed at the correct temperature. Usually this requires placing the containers holding the buffers into a water bath - ideally thermostatically controlled, but you can just about get away with adjusting the temperature with hot or cold water until your buffers are at their correct temperatures. Only then can you proceed to calibrate the meter - otherwise, you're actually de-calibrating it.
- orlando
- So far gone I'm on the way back again!
- Posts: 7201
- Joined: Thu Nov 17, 2011 3:22 pm
- Location: North Norfolk: Nearest breweries All Day Brewery, Salle. Panther, Reepham. Yetman's, Holt
Re: Electronic ph meter
Dave S wrote:Does anyone know where to get buffer solution at a reasonable price? Mine cost just short of a tenner for a 250 ml 7,4 pair and they don't tend to last very long.
I've got this, they also do a pH 4 bottle too. The two cost £8 a piece delivered.
I am "The Little Red Brooster"
Fermenting:
Conditioning:
Drinking: Southwold Again,
Up Next: John Barleycorn (Barley Wine)
Planning: Winter drinking Beer
Fermenting:
Conditioning:
Drinking: Southwold Again,
Up Next: John Barleycorn (Barley Wine)
Planning: Winter drinking Beer
-
- Even further under the Table
- Posts: 2514
- Joined: Sun Apr 01, 2012 5:38 pm
- Location: Wirral, Merseyside
Re: Electronic ph meter
That's even dearer than mine, Orlando. I got these also from Amazon.orlando wrote:Dave S wrote:Does anyone know where to get buffer solution at a reasonable price? Mine cost just short of a tenner for a 250 ml 7,4 pair and they don't tend to last very long.
I've got this, they also do a pH 4 bottle too. The two cost £8 a piece delivered.
Best wishes
Dave
Dave
- orlando
- So far gone I'm on the way back again!
- Posts: 7201
- Joined: Thu Nov 17, 2011 3:22 pm
- Location: North Norfolk: Nearest breweries All Day Brewery, Salle. Panther, Reepham. Yetman's, Holt
Re: Electronic ph meter
Apologies I didn't read the post properly and missed you had put for the pair, I thought you meant a tenner each.
I am "The Little Red Brooster"
Fermenting:
Conditioning:
Drinking: Southwold Again,
Up Next: John Barleycorn (Barley Wine)
Planning: Winter drinking Beer
Fermenting:
Conditioning:
Drinking: Southwold Again,
Up Next: John Barleycorn (Barley Wine)
Planning: Winter drinking Beer
-
- Even further under the Table
- Posts: 2514
- Joined: Sun Apr 01, 2012 5:38 pm
- Location: Wirral, Merseyside
Re: Electronic ph meter
There must be someone doing them cheaper. I've had 6 at most calibrations out of mine and they need replenishing. Gets to be a dear do, this brewing lark.orlando wrote:Apologies I didn't read the post properly and missed you had put for the pair, I thought you meant a tenner each.
Best wishes
Dave
Dave
-
- Even further under the Table
- Posts: 2514
- Joined: Sun Apr 01, 2012 5:38 pm
- Location: Wirral, Merseyside
Re: Electronic ph meter
Interestingly, I've just bought a copy of my very first brewing book, 'Brewing Better Beers' by Ken Shales. Mostly out of nostalgia really, as it was only 1p + del. Having had a quick skim through I don't think there's a single mention of pH
.

Best wishes
Dave
Dave
- gregorach
- Under the Table
- Posts: 1912
- Joined: Wed Jun 08, 2011 10:07 am
- Location: Edinburgh
- Contact:
Re: Electronic ph meter
It's been a while, I can't actually remember.Hogarth wrote:How inaccurate have you found the PH strips to be, Dunc?gregorach wrote:Yeah, I've become extremely sceptical about the value of pH strips since buying a decent meter.
Cheers
Dunc
Dunc
- gregorach
- Under the Table
- Posts: 1912
- Joined: Wed Jun 08, 2011 10:07 am
- Location: Edinburgh
- Contact:
Re: Electronic ph meter
Nah, the temperature correction in the meter is to correct the response of the electrode, which also varies with temperature.greenxpaddy wrote:Flipping heck. The manual didnt say that. The other thing was it said it recognises the 7.0ph solution on calibration. I then thought it would measure temp and calibrate it knowing the buffer temp versus accurate ph7.0. That is clearly possible....gregorach wrote:Yeah, but the buffers themselves are only accurate at the correct temperature, and they're surprisingly sensitive to variations. Your typical 7.01 pH buffer is only actually 7.01 pH at exactly 25 degrees (or whatever it states on the bottle). The temperature correction for the meter is an entirely separate matter. Calibration must be performed at the correct temperature. Usually this requires placing the containers holding the buffers into a water bath - ideally thermostatically controlled, but you can just about get away with adjusting the temperature with hot or cold water until your buffers are at their correct temperatures. Only then can you proceed to calibrate the meter - otherwise, you're actually de-calibrating it.
Cheers
Dunc
Dunc
- Eric
- Even further under the Table
- Posts: 2920
- Joined: Fri Mar 13, 2009 1:18 am
- Location: Sunderland.
Re: Electronic ph meter
How accurate need we be Dunc? Better than a point or two?gregorach wrote:It's been a while, I can't actually remember.Hogarth wrote:How inaccurate have you found the PH strips to be, Dunc?gregorach wrote:Yeah, I've become extremely sceptical about the value of pH strips since buying a decent meter.
I don't have a meter and so can't validate an argument, but have we advanced beyond balancing buffers to enable the brew to do what it can do best?
Now I've read that pH is influenced by temperature and that of buffer solutions will vary with temperature, but that variation is in some way proportional to the pH, such that it is my belief that any variance isn't really significant over the range of temperatures and pH in brewing and correction is necessary only for the instrument in use. But why calibrate at pH 7 when brewing starts with a mash that is roughly fifty times more acidic than that level? To me that would be similar to calibrating your mash thermometer at either 17,000 or minus 205 degrees centigrade.gregorach wrote:Nah, the temperature correction in the meter is to correct the response of the electrode, which also varies with temperature.greenxpaddy wrote:Flipping heck. The manual didnt say that. The other thing was it said it recognises the 7.0ph solution on calibration. I then thought it would measure temp and calibrate it knowing the buffer temp versus accurate ph7.0. That is clearly possible....gregorach wrote:Yeah, but the buffers themselves are only accurate at the correct temperature, and they're surprisingly sensitive to variations. Your typical 7.01 pH buffer is only actually 7.01 pH at exactly 25 degrees (or whatever it states on the bottle). The temperature correction for the meter is an entirely separate matter. Calibration must be performed at the correct temperature. Usually this requires placing the containers holding the buffers into a water bath - ideally thermostatically controlled, but you can just about get away with adjusting the temperature with hot or cold water until your buffers are at their correct temperatures. Only then can you proceed to calibrate the meter - otherwise, you're actually de-calibrating it.
In 1909 Sorensen introduced p[H] to the world after using a galvanometer in place of colour sensitive strips to measure the ion concentration of enzyme activity in proteins in the Carlsberg Laboratory. I wonder if his purpose and methods were intended for the advancement of brewing? A few years later institutions took his work and changed his definition, adapting it for pure science such that I wonder if that is why today brewers struggle to know what it means?
Certainly I'm baffled.
Without patience, life becomes difficult and the sooner it's finished, the better.
- gregorach
- Under the Table
- Posts: 1912
- Joined: Wed Jun 08, 2011 10:07 am
- Location: Edinburgh
- Contact:
Re: Electronic ph meter
As as been observed elsewhere, I'm a bit OCD when it come to numbers... But the inaccuracy I noticed in my pH strips was certainly of the order of several points. I just can't remember exactly how many. Could even be up to +/- 0.5.Eric wrote:How accurate need we be Dunc? Better than a point or two?
Because 7.01 and 4.01 are the standard calibration points for all pH meters. Those are the calibration buffers you can buy most easily, and they're the values the meter is expecting when you put it into calibration mode.gregorach wrote:But why calibrate at pH 7 when brewing starts with a mash that is roughly fifty times more acidic than that level?
Cheers
Dunc
Dunc