Do pH probes 'go off'?
About 18 months ago I bought a Voltcraft PH-100ATC Ph meter.
A couple of days ago was brewday. During the first 15 minutes of the mash I calibrated (or tried to) the pH meter using pH 4 and pH 7 buffers. New ones, by the way.
The pH 7 point was wandering all over the place every time the probe was put back into it. By anything up to +0.6 -0.3. This meant the slope from the pH 4 had to be constantly altered. I must have gone back and forth between the two buffers about 8 times.
Eventually I was happy with the readings at 4 and 7. Dropped the probe into some cooled wort and got 5.2. No idea if this was real or not!!
Some details;
the probe is stored in 'Growth Technology' storage solution The buffers are 'Growth Technology The probe is washed in reverse osmosis water between readings Any thoughts, please?
Guy
A couple of days ago was brewday. During the first 15 minutes of the mash I calibrated (or tried to) the pH meter using pH 4 and pH 7 buffers. New ones, by the way.
The pH 7 point was wandering all over the place every time the probe was put back into it. By anything up to +0.6 -0.3. This meant the slope from the pH 4 had to be constantly altered. I must have gone back and forth between the two buffers about 8 times.
Eventually I was happy with the readings at 4 and 7. Dropped the probe into some cooled wort and got 5.2. No idea if this was real or not!!
Some details;
the probe is stored in 'Growth Technology' storage solution The buffers are 'Growth Technology The probe is washed in reverse osmosis water between readings Any thoughts, please?
Guy
-
- Falling off the Barstool
- Posts: 3899
- Joined: Thu Jan 24, 2013 11:07 pm
- Location: South Wales UK.
Re: Do pH probes 'go off'?
Guy,
I thought it was lovely sitting in the garden drinking copious amounts of Beer today!
WA
Sent from my garden.
-
- Hollow Legs
- Posts: 374
- Joined: Tue Jun 05, 2018 9:19 am
Re: Do pH probes 'go off'?
Might need a new probe if they’re not expensive? Contact the company for spares...
- Eric
- Even further under the Table
- Posts: 2879
- Joined: Fri Mar 13, 2009 1:18 am
- Location: Sunderland.
Re: Do pH probes 'go off'?
They don't last for ever. My first probe didn't last long, but wasn't giving erratic readings, instead it just continued to drift.
For readings to be practical and meaningful they should be taken with the sample and probe at very similar temperatures, close to 20C and constant. For our purpose perfect precision isn't essential, but we should avoid taking potentially misleading readings. The following is my procedure. The probe is rinsed in DI water before and after every stage and excess surplus DI water gently shaken off the probe.
With DI water, probe and sample at the same and right temperature, the sample is gently stirred using the probe. The probe is left in the sample and allowed to settle while the reading observed. This is repeated until the reading stabilises quicky and remain constant for about a minute, then recorded. The rate at which this happens depends upon the buffering power of the medium being measured. Buffers used to calibrate the meter will settle to a constant reading more quickly than will, for example, low gravity wort runnings from a thin mash created using RO liquor. DI water itself has virtually no buffering power and therefore a large range of potential pH.
My first probe was replaced when the pH reading began stabilising, but continued to rise very slowly for many minutes. I replaced it with one of these, although maybe not this particular supplier. I found this better than the original in that it has lasted much longer and used for vastly more readings. It also can be opened to allow the calibration fluid to flow into the membrane and be topped up using the syringe supplied, something I was unable to do with the original.
As your readings were erratic Guy, I wonder if there might be an intermittent contact in the wiring. Also, as no doubt you have found, any reading can be disturbed if the body is moved when taking a reading. You can get pH probe cleaning fluid, but it will likely cost more than the probe above.
For readings to be practical and meaningful they should be taken with the sample and probe at very similar temperatures, close to 20C and constant. For our purpose perfect precision isn't essential, but we should avoid taking potentially misleading readings. The following is my procedure. The probe is rinsed in DI water before and after every stage and excess surplus DI water gently shaken off the probe.
With DI water, probe and sample at the same and right temperature, the sample is gently stirred using the probe. The probe is left in the sample and allowed to settle while the reading observed. This is repeated until the reading stabilises quicky and remain constant for about a minute, then recorded. The rate at which this happens depends upon the buffering power of the medium being measured. Buffers used to calibrate the meter will settle to a constant reading more quickly than will, for example, low gravity wort runnings from a thin mash created using RO liquor. DI water itself has virtually no buffering power and therefore a large range of potential pH.
My first probe was replaced when the pH reading began stabilising, but continued to rise very slowly for many minutes. I replaced it with one of these, although maybe not this particular supplier. I found this better than the original in that it has lasted much longer and used for vastly more readings. It also can be opened to allow the calibration fluid to flow into the membrane and be topped up using the syringe supplied, something I was unable to do with the original.
As your readings were erratic Guy, I wonder if there might be an intermittent contact in the wiring. Also, as no doubt you have found, any reading can be disturbed if the body is moved when taking a reading. You can get pH probe cleaning fluid, but it will likely cost more than the probe above.
Without patience, life becomes difficult and the sooner it's finished, the better.
Re: Do pH probes 'go off'?
I had a similar situation with my rather expensive Milwaukee ph meter ( a good £60-£80 I think) worked fine for a couple years, always cleaned and stored the probe in storage solution. It wouldn’t calibrate at all and a new probe was more than a new one!
So I bought one of these and it’s brilliant. Best part is when it dies I’ve only lost £10
So I bought one of these and it’s brilliant. Best part is when it dies I’ve only lost £10
Cheers and gone,
Mozza
Mozza
- Eric
- Even further under the Table
- Posts: 2879
- Joined: Fri Mar 13, 2009 1:18 am
- Location: Sunderland.
Re: Do pH probes 'go off'?
I've often wondered about buying a cheaper one, particularly since finding the cheap probe to be at least equal to the much more expensive one. How frequently do you calibrate yours and well does it hold calibration?mozza wrote: ↑Sat Jul 13, 2019 3:37 pmI had a similar situation with my rather expensive Milwaukee ph meter ( a good £60-£80 I think) worked fine for a couple years, always cleaned and stored the probe in storage solution. It wouldn’t calibrate at all and a new probe was more than a new one!
So I bought one of these and it’s brilliant. Best part is when it dies I’ve only lost £10
Without patience, life becomes difficult and the sooner it's finished, the better.
-
- Even further under the Table
- Posts: 2653
- Joined: Mon Mar 22, 2010 7:10 pm
- Location: Christchurch, Dorset
Re: Do pH probes 'go off'?
Thanks. When I first got the meter it calibrated easily without having to go back and forth between the buffers. I've got another probe in the drawer. A fairly cheap one from ebay. Sounds like it might be worth trying it out.
Guy
Guy
Re: Do pH probes 'go off'?
I don’t brew that regularly so I always calibrate it, but to be honest it only needs a slight tweak every now and then. I’m really impressed by the accuracy of itEric wrote:I've often wondered about buying a cheaper one, particularly since finding the cheap probe to be at least equal to the much more expensive one. How frequently do you calibrate yours and well does it hold calibration?mozza wrote: ↑Sat Jul 13, 2019 3:37 pmI had a similar situation with my rather expensive Milwaukee ph meter ( a good £60-£80 I think) worked fine for a couple years, always cleaned and stored the probe in storage solution. It wouldn’t calibrate at all and a new probe was more than a new one!
So I bought one of these and it’s brilliant. Best part is when it dies I’ve only lost £10
Cheers and gone,
Mozza
Mozza
-
- Drunk as a Skunk
- Posts: 904
- Joined: Tue Jan 15, 2013 11:42 am
- Location: Newton-le-Willows ( St Helens/Warrington)
Re: Do pH probes 'go off'?
Probes do go off, something you might try is soaking your probe in a warm solution of Biological detergent (for the washing machine). The enzymes may remove any proteins that are blocking the poures in the glass membrane.
It works sometimes.
Atb. Aamcle
It works sometimes.
Atb. Aamcle
Re: Do pH probes 'go off'?
Eric wrote:
I've often wondered about buying a cheaper one, particularly since finding the cheap probe to be at least equal to the much more expensive one. How frequently do you calibrate yours and well does it hold calibration?
First time out the box in a couple of months and virtually no calibration needed.
Well worth a tenner even if it’s there as a spare
Cheers and gone,
Mozza
Mozza
- Eric
- Even further under the Table
- Posts: 2879
- Joined: Fri Mar 13, 2009 1:18 am
- Location: Sunderland.
Re: Do pH probes 'go off'?
That seems to be so. It would be interesting to compare the linearity between and outside the calibration points for both machines.
Mine tends to be used in flourishes. Often I won't bother when time is short or for a basic brew, but it gets a bashing on occasions when several readings are taken during mash and sparge, plus start and finish of the boil. Measuring pH of other peoples beer is sometimes interesting, gets used when making invert sugars and is invaluable when diluting acids to particular strengths.
Busy working on the house at present and the garage, where I brew, is a right mess. As one doesn't take up much space I'll order one soon.